gabrojas Posted August 24, 2022 at 04:58 PM Posted August 24, 2022 at 04:58 PM (edited) Hallo, gibt es eine Möglichkeit die automatische Kalibrierfunktion (ASC) des Sensirion SCD30 CO2 Sensors zu deaktivieren? Wir nutzen 12 dieser Sensoren für bauphysikalische Experimente und würden ASC gerne deaktivieren, aber finde dazu nichts in der API Beschreibung. Danke im voraus...Gabriel Edited August 24, 2022 at 05:27 PM by gabrojas 1 Quote
Superp Posted August 26, 2022 at 06:37 AM Posted August 26, 2022 at 06:37 AM Cross reference: CO2 Bricklet 2.0 accuracy Quote
gabrojas Posted August 26, 2022 at 10:09 AM Author Posted August 26, 2022 at 10:09 AM Thanks! That helps already a lot, I had read it before, but after reading it again I still don't know if I can switch it off via API or if I have to change sensor firmware. A short answer on that would be highly appreciated, then I would know if feasible for me or if I need to involve collegues...Thanks Quote
Superp Posted September 1, 2022 at 08:26 PM Posted September 1, 2022 at 08:26 PM Hi Gabrojas, It is not possible to switch ASC on/off via API with the current firmware. That other thread ultimately led to firmware 2.0.3, which forces ASC on at startup. Before that, at least some CO2 Bricklets erroneously shipped with ASC switched off. I would welcome the option to control ASC manually, but if you know how to build firmware, it might be feasible to implement yourself. This is the diff for the above FW change. Hope this helps. Quote
gabrojas Posted September 22, 2022 at 10:18 AM Author Posted September 22, 2022 at 10:18 AM Thanks Superp, Do you know: does that mean that if I install firmware 2.0.2 the ASC will be turned off? The changelog for 2.0.3 says "Make sure ASC is enabled on each startup", but does that mean that ASC is disabled with 2.0.2 after a startup or not necessarily and if ASC is activated (now with 2.0.3) it stays activated, even if you flash with 2.0.2? Quote
Superp Posted October 1, 2022 at 08:00 AM Posted October 1, 2022 at 08:00 AM I don't know. Did you read the thread I linked to above? Firmware < 2.0.3 did not touch ASC, it was assumed to be activated during manufacturing. But, apparently, at least some Bricklets were shipped with ASC not activated. So, I would say, a Bricklet that has never run 2.0.3 has ASC in an unknown state. 2.0.3. activates ASC. 2.0.2. and older do not touch ASC. So the question is: does the Bricklet retain the state of ASC between restarts? 1 Quote
gabrojas Posted October 7, 2022 at 01:51 PM Author Posted October 7, 2022 at 01:51 PM Thanks! Yes, I read the thread and I read the GitHub documentation (and the available changelog. But it wasn't conclusive for me how some sensors with old firmware, where shipped with and some without ASC turned off. But thanks to your clarification I get it now. But what you say points to the conclusion that the bricklet remains the state of ASC between restarts...Seems like there is no way around reprogramming the firmware. Thanks again! Quote
Superp Posted October 13, 2022 at 08:57 PM Posted October 13, 2022 at 08:57 PM I think you're right – once a Bricklet has booted 2.0.3, ASC is on forever – but I am not sure. Would be nice if e.g. @borg could confirm. Quote
borg Posted October 14, 2022 at 09:49 AM Posted October 14, 2022 at 09:49 AM vor 12 Stunden schrieb Superp: I think you're right – once a Bricklet has booted 2.0.3, ASC is on forever – but I am not sure. Would be nice if e.g. @borg could confirm. You are correct. The SCD30 documentation states: Zitat ASC status is saved in non-volatile memory. When the sensor is powered down while ASC is activated SCD30 will continue with automatic self-calibration after repowering without sending the command. 1 Quote